Help with Analog signal

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I am attempting to setup a Varian 3400 GC with FID detector.
I have successfully plotted the signal on an old chart recorder, and the signal looks clean. However, I would like to hook it up to an Shimadzu CR5A Integrator (from an old HPLC). However, when I attach it to the integrator, I get a constantly increasing baseline (nice straight line at about a 45 degree angle) and I don't seem to be getting any actual signal.

Any suggestions? Is this setup even possible?
Possibilities :

Cable connection to CR-5A is wrong , so you are getting blank signal , increasing during warming up of the integrator.

The attenuation is very low , so you are seeing the very small changes on the signal.

Change the hydrogen flow a little bit and observe the signal if a you see a sudden change , it gives you some idea about the signal.
2 posts Page 1 of 1

Who is online

In total there are 33 users online :: 1 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Semrush [Bot] and 32 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry