Low pressure+1100 LC+six port valve of triple quad
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
2 posts
Page 1 of 1
There is a drop in cap pressure after valve switch from detector to load; the starting pressure is 60 bar but after valve switch it stablizes at 10-12 bar. What are the possible causes? I have made sure there are no leaks!
-
- Posts: 66
- Joined: Sun Oct 17, 2010 9:34 pm
Is this an Agilent triple quad? The rotor seal and/or stator face in the MS switching valve may need replacing or there could be a blockage in the waste line. You can remove the switching valve and try giving all the bits in there a clean to get a bit more life out of it.
2 posts
Page 1 of 1
Who is online
In total there are 107 users online :: 0 registered, 0 hidden and 107 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 107 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 107 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science