negative intercept in calibration curve

Discussions about GC and other "gas phase" separation techniques.

18 posts Page 1 of 2
Hi everybody,

I am having negative intercepts for some of my analytes I injected on GC-uECD. I am wondering if I should force the origin to zero or just add the absolute value of intercept in the calculation?


regards,
kaka
More details are required. The negative intercept can indicate real things going on in the chromatographic system - in which case it is telling you things that can not be ignored - or it may be the result of noise - in which case the better option is reconsideration of calibration levels and number of injections at those levels. (Or the difference actually is so small that deviation from zero can be ignored.)

How are you making up standards and in what matrix? What kinds of compounds are you running? Are you using an internal standard?
More details are required. The negative intercept can indicate real things going on in the chromatographic system - in which case it is telling you things that can not be ignored - or it may be the result of noise - in which case the better option is reconsideration of calibration levels and number of injections at those levels. (Or the difference actually is so small that deviation from zero can be ignored.)

How are you making up standards and in what matrix? What kinds of compounds are you running? Are you using an internal standard?
Thanks for your reply. Standards are hexachlorobenzene, PCB101, penta/hexa-chloroethanes and mirex and they are in isooctane. I made 5 levels by serial dilution spanning three order of magnitude. The intercept value is significant and not near to zero.
Hi Kaka

An ECD is not quite linear in response, if you are calibrating over a wide range, or using high concentrations and then extrapolating below the range for which you have actual measurements this might be why you get a negative intercept.

Peter
Peter Apps
Hi Kaka

An ECD is not quite linear in response, if you are calibrating over a wide range, or using high concentrations and then extrapolating below the range for which you have actual measurements this might be why you get a negative intercept.

Peter
Hi Peter,

I see your point but the regression among different levels is pretty linear (R2= 0.99) and my sample points fall between the calibration levels. So, in this case, should I ignore intercept?

best,
Kaka
Hi Kaka

Just to be certain - you are talking about the intercept on the Y axis ? So you have a positive intercept on the X-axis.

What is the calibration equation ?

Peter
Peter Apps
Hi Kaka

Just to be certain - you are talking about the intercept on the Y axis ? So you have a positive intercept on the X-axis.

What is the calibration equation ?

Peter
Yes, it is Y intercept (value is - 2E+09). The equation for pentachloroethane is:

y=(3E+11)x - 2E+09


Kaka
So the intercept is 2/3 of a percent of the slope.

What are x and y for your lowest calibration point ?

Peter
Peter Apps
So the intercept is 2/3 of a percent of the slope.

What are x and y for your lowest calibration point ?

Peter
For lowest point, x is 3.68E-02 and y is 8.4E+09.
The intercept is too big in comparison to your data points to just ignore it. You need to trouble-shoot the analysis; either your calibration standards have lower concentrations than you think that they have, or you have a constant bias in your instrumental response.

Peter
Peter Apps
The intercept is too big in comparison to your data points to just ignore it. You need to trouble-shoot the analysis; either your calibration standards have lower concentrations than you think that they have, or you have a constant bias in your instrumental response.

Peter
Thanks but I got negative intercept with this compound only and not for others. I injected them as a mix with equal concentrations.
If the other compounds are OK then there is probably nothing wrong with the instrument. How do you make up you standards ?, at what stage do you have all the compounds in solution together ?

Peter
Peter Apps
If the other compounds are OK then there is probably nothing wrong with the instrument. How do you make up you standards ?, at what stage do you have all the compounds in solution together ?

Peter
right from the beginning. I made a stock solution containing the mix of 5 components and then the stock solution was serially diluted to 5 levels.
If the other compounds are OK then there is probably nothing wrong with the instrument. How do you make up you standards ?, at what stage do you have all the compounds in solution together ?

Peter
right from the beginning. I made a stock solution containing the mix of 5 components and then the stock solution was serially diluted to 5 levels.
Step one is to make the stock solution again - by far the most likely explanation is operator errror at that stage.

Peter
Peter Apps
Do you use an internal standard? And if so, the same internal standard for all compounds? And does an injection of solvents with no added internal standard or analyte give any peak - particularly in the location for the internal standard?
18 posts Page 1 of 2

Who is online

In total there are 33 users online :: 0 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry