Rising curve at the end of chromatogram
Posted: Mon Nov 19, 2012 2:31 am
by ito_miura
Hi all.
I have a problem where the curve at the end of the chromatogram is rising instead of flat, near the base peaks. Is this a normal occurence? Otherwise, why could possibly make the chromatogram spike up?
Thanks in advance.
Re: Rising curve at the end of chromatogram
Posted: Mon Nov 19, 2012 3:58 am
by DR
details please.
Posted: Mon Nov 19, 2012 6:49 pm
by Bigbear
Yes details.
If it is what I think, yes its normal. Most columns give a slightly rising baseline due to the temp. ramp. It looks worse if there is no major peaks to normalize on.
Posted: Tue Nov 20, 2012 7:32 am
by ito_miura
dear all,
so it's normal if there's no peak anymore in the end analysis the baseline is rising up?
Posted: Tue Nov 20, 2012 7:48 am
by Peter Apps
Let's try to get just one very basic piece of information - is this gas chromatography or liquid chromatography ?
Posted: Tue Nov 20, 2012 2:24 pm
by ito_miura
dear peter,
it is gas chromatrography with mass spectrometry and also pyrolysis...
Posted: Tue Nov 20, 2012 2:37 pm
by Peter Apps
Then the rising baseline is normal (presuming that you are doing temperature programming) and is usually due to column bleed and the elution of unresolved heavy compounds.
Posted: Wed Nov 21, 2012 2:58 am
by ito_miura
okey thank you if this is normally happen......
have a nice day....