0.53 precolumn on a DB-5 0.32 column

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I am doing a determination of waxes in oils with 10 meters of DB-5 0.32 column with an on column injector.

I was wondering if adding a 0.53 deactivated fused silica precolumn (to facilitate the use of larger diameter syringers for the on column injection) in front of the column could interfere with the analysis or the way the sample goes into the column.

Could you think of anything against doing this analytically-wise?
very commonly done procedure.

You must be careful when connecting the precolumn to the column.

No exposed edges, etc.

A deactivated union with ferrules, butting up the columns to a small deactivated void would ensure an enduring seal during thermal stresses.

But I would do it in a heartbeat.

Happy Thanksgiving to all, and to all: stuffing, gravy, and your favorite meat dish.

Rod
2 posts Page 1 of 1

Who is online

In total there are 33 users online :: 0 registered, 0 hidden and 33 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 33 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry