Our lab recently purchased a waters PAD detector, attached to an agilent. Anyone get this set up to work correctly? We have signal, but can not reproduce standards for sugar alcohols. I am thinking maybe it is the ED settings. Let me know if anyone has any experience with this.

Thanks!
Before you start talking about settings on the PAD, the LC system must be spotlessly clean and passivated to start. After that, then all the glassware, solvents, columns and need to be clean.

After that your noise should be lower and then you can start talking about settings.
It was a brand new system so it was clean. Turned out it was the pump making everything elute late. It was only delivering 1/3 of the volume it should have with the column in line.
3 posts Page 1 of 1

Who is online

In total there are 68 users online :: 4 registered, 0 hidden and 64 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Baidu [Spider], Semrush [Bot] and 64 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry