Helium %age determination in natural gas (Methane)

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
I am using capillary column and tcd detector H2 as a carrier but He peak is not coming only methane (N2 and Oxygen due to bad sample taking).
Oh, the helium is coming out. But if you can see it depends upon the amount the capillary will separate from the helium peak from the valve upset and the O2 and N2 peak.

Plus the peak is probably negative in height.

Try using a more appropriate carrier gas and column.

best wishes,

Rod
Thanks for replying.
What you prefer for the carrier gas.
I think column is ok. It is seperating all the peaks at -15'C. And I have only one capillary column(plot q). So no option for column.
To determine hydrogen only, nitrogen is best, argon is second best.

I have no idea what column you are referring to in (TQ-300). Nothing is found when searching the internet.

best wishes,

Rod
One notation:

You can afford to cool a GC oven at -15C but can't afford a second column for $100?

Different strokes for different folks.

Rod
5 posts Page 1 of 1

Who is online

In total there are 32 users online :: 1 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Semrush [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry