I second that notion: "I highly suggest you read the manual that was provided."
Once you learn how to use the basic software functionality to add components to your processing method you can use a feature called 'RT Reference' in the components tab as an aid to deal with RT shift.
First, you'll need a peak which appears in every chromatogram and is as isolated from the others as possible.
Next, you'll want to open the RT Window for this referenced peak large enough that it is labeled in every chromatogram (hence the need for it to be isolated from the others).
Lastly, assign this component as the 'RT Reference' for the others. As it shifts in RT, the component labels for all the other peaks will shift accordingly without having to assign any special Peak Match rules or adjust any RT Windows.
If you do not understand the basics then these instructions are probably very confusing to you. It would be wise to take the time to familiarize yourself with the manual. Or better yet, take an online course given by Waters. Here is one that would cost $100 that you can take on the web:
http://www.waters.com/waters/educationI ... =134718981
Good luck!