Is it good practice to purge the injector before every sample set is run?
yes..
What does the purge injector actually do? does it matter what solvent is running through the system when this is being carried out for example water or mobile phase and is anybody aware of any literature on how it works.
What does the purge injector actually do? does it matter what solvent is running through the system when this is being carried out for example water or mobile phase and is anybody aware of any literature on how it works.
Purge injector washes your injector with your selected solvent. It is required use water for purging when you shift for a high buffer to high organic mobile phase to avoid salt precipitation. Then you have to purge with your mobile phase. Or shifting from normal phase to reversed phase (or vice versa) you have to use an intermediate polarity solvent first to avoid phase formation.

You especially need injector wash; when solvent is changed, bubles appears on syringe, system fills with air etc. That's why it is a good practice to purge injector frequently..
4 posts Page 1 of 1

Who is online

In total there are 5 users online :: 1 registered, 0 hidden and 4 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 4 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry