Pinnacle DB C18 lifetime for opiate analysis

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
We're getting ready to do a study on optimum lifetime for Pinnacle DB C18 3um 100x2.1mm columns used in urine analysis for opiates. It looks like Restek is loathe to give any estimations on column life so does anyone have experience with column life in urine analysis or know a good place for background research?
Of course, Restek will give you no estimation of the column lifetime. It depends heavily on the method (mobile phase, temperature, gradient or isocratic, etc.) and the kind of samples run (which matrix, how much of it, etc.). And the third factor is luck!
You may get several thousands of injections out of a column, or the very same column may die after just a few.
Urine samples, as any samples of biological origin, are not the easist stuff for a column. I'd strongly suggest to use a guard column to protect your main analytical column.
jsupica-

I'm interested in hearing more about your study, perhaps even collaborating.

Can you tell me what industry you are in?
3 posts Page 1 of 1

Who is online

In total there are 29 users online :: 1 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry