Retention time shift problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
The retention of my internal standard and target both shifted back 40sec. I am curious what may caused the problem. I first though it maybe I mis preparation of my mobile phase, but after I re-prepare it, the retention time are still the same, 40 sec later than before.

This just happened during i was running the standard curve.The first few standards are just fine, having the same retention time as previous runs. But from the 4th standard, it suddenly changed, and all for the others.

I understand the column efficiency may decrease gradually, but why this just happened suddenly and keeps stable afterwards?
Isocratic or gradient method? If isocratic, pre-mix or on-line mixing of the mobile phase?

Did both peaks shift by exactly the same amount, or did they shift by the same percentage?

And finally, did your dead time ("solvent front") also shift (and if so, how much)?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry