Dissolution Braket Standard Repeatability Fail
Posted: Fri Nov 02, 2012 9:41 pm
Hi All,
During the validation of a new method, based on USP, I have been observing a reccurent, (but intermitent) problem. After the system suitability passes (precision NMT 2.0% Tailing NMT 2.0, Plates NLT 1,800, Accuracy (98-102%), the second or third standard braket drops about 3% of its area (limit 2%), and then after it remains during the rest of the run with the same dropped value
. It had happen with two different Agilent HPLC systems with different detectors (VWD and DAD), as well different analysts, and with different standard concentration (in the linear range)
The method HPLC parameters are:
Column: 125 x 4.0 mm x 5µm (Discvery C18)
Mobile Phase: Methanol:Water:Acetic acid (80:20:0.5)
Flow rate: 2.0 mL/min
Temperature: 40C
Detection UV @ 353nm
injection 20µL
RT ~2 min
samples and standard are diluted with pH 8.0 boric acid buffer
Thanks in advance for any help to solve this apparent mistery
During the validation of a new method, based on USP, I have been observing a reccurent, (but intermitent) problem. After the system suitability passes (precision NMT 2.0% Tailing NMT 2.0, Plates NLT 1,800, Accuracy (98-102%), the second or third standard braket drops about 3% of its area (limit 2%), and then after it remains during the rest of the run with the same dropped value

The method HPLC parameters are:
Column: 125 x 4.0 mm x 5µm (Discvery C18)
Mobile Phase: Methanol:Water:Acetic acid (80:20:0.5)
Flow rate: 2.0 mL/min
Temperature: 40C
Detection UV @ 353nm
injection 20µL
RT ~2 min
samples and standard are diluted with pH 8.0 boric acid buffer
Thanks in advance for any help to solve this apparent mistery