I am using an Waters XBridge BEH C18 Column. The test compounds I work with elute with the solvent front at 98% Aqueous Mobile Phase
Mobile Phase A: Water (0.1%/0.05% A-/HA Formate Buffer)
Mobile Phase B: ACN (0.1%/0.05% A-/HA Formate Buffer)
Most of my attempts have been running the system at 0.75 - 1.5 mL/min.
I was wondering if there were any options to gaining retention using this column. I'm trying a round of buffers to see if this effects the system, but I don't suspect it will (no readily labile hydrogens on these compounds).
Will trying a different mobile phases, changing flow rate, temperature, ect really effect retention or should I simply look into getting a new column (and if so, do you have any recommendations?) I'm relatively new to LC method development and been trying to read up as much as I can, but could certainly use some guidence.
I've provided a link for one of the compounds I am looking at below.
http://www.sigmaaldrich.com/catalog/pro ... ®ion=US