There are a number of possibilities - some good some bad. And more information is needed to sort it out.
Having said that, with your second post, the next question that comes to mind is could this be a matter of peaks shifting as a result of column overload from an earler peak in the chromatogram? I would expect this to happen in samples but not a solution standard. And it woud not have to happen in all samples, just those whith a large matrix component that elutes just before the analytes of interst. Alternatively, the samples could leave garbage on the column that affects the chromatography, but that garbage goes away after the instrument stands for a while.
So, what does a total ion chromatogram for the samples look like? Is there a large peak that elutes just ahead of the shifted peaks? (This requires full scan acquisition to detect the extra compund.)
The standard at the end of the sequence give a clue. A shifted retention time would indicate a change to the instrument; a normal retention time suggests the problem is with individual samples -- like peak overloading mentioned above.
On a long sequence, several standards through the sequence are a good idea. When instrument operation becomes questionable, all results back to the last known good sample (a sample with a known result that ran correctly) are questionable. And if you the last sample you ran was just before all the samples in the run, you get to run them all again if you need to be sure of results. And the multiple control samples or standards would help to confim what you are seeing in the standard at the end of the sequence.