During the validation of a new method, based on USP, I have been observing a reccurent, (but intermitent) problem. After the system suitability passes (precision NMT 2.0% Tailing NMT 2.0, Plates NLT 1,800, Accuracy (98-102%), the second or third standard braket drops about 3% of its area (limit 2%), and then after it remains during the rest of the run with the same dropped value

The method HPLC parameters are:
Column: 125 x 4.0 mm x 5µm (Discvery C18)
Mobile Phase: Methanol:Water:Acetic acid (80:20:0.5)
Flow rate: 2.0 mL/min
Temperature: 40C
Detection UV @ 353nm
injection 20µL
RT ~2 min
samples and standard are diluted with pH 8.0 boric acid buffer
Thanks in advance for any help to solve this apparent mistery