MSTOP error msg

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
I installed MSD Chemstation version E.02.00 with 5973N and 6890N, after configuring the instrument and when I start instrument 1 am getting this error msg "unable to start MSTOP! Run productivity tool/config, and delete and then reconfigure instrument!".

Can anyone suggest what the problem could be.

Thanks
Shyam
I installed MSD Chemstation version E.02.00 with 5973N and 6890N, after configuring the instrument and when I start instrument 1 am getting this error msg "unable to start MSTOP! Run productivity tool/config, and delete and then reconfigure instrument!".

Can anyone suggest what the problem could be.

Thanks
Shyam
The Chemstation software is corrupt. I have seen that several times with the E revision. Go into the config editor, click on instrument 1 (or whichever installation you are having issues with if you have more than one) and click through the menus. You don't need to make any changes. Once you have clicked through all the way it will ask if you want to update the configuration. Answer yes and let it finish. Close the config editor and you should be able to open Chemstation normally. If that doesn't work you will need to un-install, then re-install the Chemstation software.
First thing Monday morning, I encountered this same issue. The only difference is my software is version D.00.01. The steps given by MX304 corrected the problem in less than 5 minutes. I did not have to un-install and re-install chemstation. I am so glad I found Chromatograpy Forum. Big thanks to MX304!

Eric
First thing Monday morning, I encountered this same issue. The only difference is my software is version D.00.01. The steps given by MX304 corrected the problem in less than 5 minutes. I did not have to un-install and re-install chemstation. I am so glad I found Chromatograpy Forum. Big thanks to MX304!

Eric
Glad I could help!
Greeting,

I'm experiencing the same error when I attempt to log-on to chemstation. I reconfigured my instrument and uninstalled the chemstation software, however, this didn't rectify the error. Can anyone offer any assistance?

Regards,

Collie
Greeting,

I'm experiencing the same error when I attempt to log-on to chemstation. I reconfigured my instrument and uninstalled the chemstation software, however, this didn't rectify the error. Can anyone offer any assistance?

Regards,

Collie
We had a similar problem upgrading and older 5973N/6890N unit from WinNT to Win7 and the E version of chemstation. What we found was that the old Jet Direct interface card in the GC was too slow to process the data coming from the Chemstation. We had to upgrade the ROM on the motherboard and the LAN interface and this solved the problem. Another instrument we upgraded would not allow the MSD to receive an IP address from the newer BootP that runs as a process, on XP I simply installed an older version of the BootP program but when upgrading to Win7 I could not get that to run so we simply run the LAN cable to another XP computer and have it give the MSD its IP address then plug it back into the Win7 Chemstation.

There is a spread sheet on the Agilent site that gives compatibilities between instrument hardware, OS version and Chemstation Versions that helps pretty well. Getting all of the parts to work together is sometimes more Voodoo than science when it comes to Chemstation it seems. We have one computer that will refuse to communicate with the GC/MS even when installed with the same WinXP disks as another computer that will work. One runs fine but will not show all compounds in data analysis on that machine or when viewed across the network on another machine, but if you copy the data file to that second machine it will display everything correctly. I think in that case I have narrowed it down to the fact the harddrives are in a Raid1 mirror, but still not positive.

If you have any spare computers or laptops try those also, maybe you will hit on a combination that works. Good Luck, I hope you can get it working.
The past is there to guide us into the future, not to dwell in.
6 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry