Recently I am trying to use LC/MS to do impurity identification of nicotine solution. According to the method that our client provided, I got noting. The first thing that concerns me is the mobile phase ion. I use 0.01%TFA as mobile phase A and 100% acetonitrile as mobile phase B according to this gradient method. I always detect peaks at 142, 163, 204 even when NO column is installed and only mobile phase is running. Then I change the mobile phases to that I used for other experiment: 0.5% FA, 2mM ammonium acetate in H20 for mobile phase A and 0.5% FA, 2mM ammonium acetate in Methanol for mobile phase B to run mobile phase only (same gradient method without column) to see the mobile phase ion. This time I still see ions at 163, and other ions are 146, 361, 361, 375 etc. I just wonder why I always see ion at 163, even for completely different mobile phase? What this ion is? Do you usually see so many ions from mobile phases at this mass range (150-400Da)? Or is there anything wrong with my instrument? (I have completely cleared both HPLC and MS source to prevent contamination). Also what are the chance for the LC/MS (ESI) to be useful for the identification of unknown despite the interfere ions from the mobile phase? What is the normal level of baseline from the mobile phase in CPS? (Say API LC/MS ESI+ mode). Thanks for any suggestions.