Nitrogen Phosphorus Detector NPD - conditioning frequency ?
Posted: Mon Oct 29, 2012 9:41 pm
I’ve been working with an NPD. I'm still trying to optimize the method and I realize that the air is consuming extremely fast (one tank in a week), so I thought that it may had a leak but I checked all the connections and did not detect any outside leak. Since I have had to turn off the gases to save gas every day and therefore the instrument I am conditioning the bead every day and I don't know if that is bad for the instrument. Another thing that I noticed is that after the bead is conditioned the chromatograms are pretty good based on base line but after about the fifth run the baseline starts getting bad (wavy and with negative peaks). Any way to improve this? Also the peak does not look clean at the tip, it actually has several tiny peaks so when I integrate the peak is actually integrating all of those little dips at the top of the peak. Do you have any suggestions, possibly? Thank you for any help. Cheers, J