Moisture hump in voa run

Posted: Mon Oct 29, 2012 7:49 pm
by sllimbri65
Image
We are experiencing an unusually large moisture hump at the start of a voa run. we use a Tekmar 3100 with an mcs loop. The hump is not there when doing a direct inject on to the gc. i don't have much experience diagnosing p&t concentrator issues but am good with a wrench. any help would be appreciated.

Re: Moisture hump in voa run

Posted: Tue Oct 30, 2012 9:51 am
by Johnny Rod
Is this a new problem, or have you always had it? Same types of samples?

Re: Moisture hump in voa run

Posted: Tue Oct 30, 2012 3:53 pm
by AICMM
sllimbri65,

Do you really need to scan down to 18? That does not solve the problem, I know but still..... If you don't scan down to 18 then, in theory, it is not stable in the quad and you will minimize the impact.

What is your split ratio?

Best regards,

AICMM

Re: Moisture hump in voa run

Posted: Tue Oct 30, 2012 8:19 pm
by sllimbri65
We scan to 35. The hump is showing ion 44 from the disolved co2 in the water. We use a 1:50 split for water samples. We have been using the same method for years.

Re: Moisture hump in voa run

Posted: Tue Oct 30, 2012 11:57 pm
by cleh
How old is your trap and which one are you using. What are your p&t parameters?

Re: Moisture hump in voa run

Posted: Wed Oct 31, 2012 12:49 am
by sllimbri65
Image
These are the concentrator parameters. The trap is a new Vocarb 3000

Re: Moisture hump in voa run

Posted: Wed Oct 31, 2012 9:35 am
by Johnny Rod
I think these have to be conditioned before use?

Re: Moisture hump in voa run

Posted: Wed Oct 31, 2012 11:21 pm
by cleh
[quoteThese are the concentrator parameters. The trap is a new Vocarb 3000
Image
Did you leak check the purge and trap after installing the trap? The ferrule on the bottom of the trap is usually graphite. I like to use graphite/vespel because it holds up better to the 250 desorb and 260 bake temps required for the Vocarb 3000 trap. If you leak check the system and you have a leak, this ferrule is usually the problem. It's hard to seat it because the bottom of the trap line is moves around. The Vocarb 3000 does need to be baked out as Johnny Rod stated. Once you have a leak tight system, you should bake out the Vocarb 3000 at 270C for at least an hour (I like to do four hours). Leak check after heating it up to 270 for ten minutes and cooling it down to make sure the ferrule has settled (you may have to tighten it a bit) then bake it out for one to four hours. Leak check again after the bake-out and you should be good to go.

Re: Moisture hump in voa run

Posted: Thu Nov 01, 2012 6:18 pm
by Bigbear
I sometimes see some wierd humps in the early part of my chromatograms on a new Vocarb 3000. They go away after a few runs. Spoke with supelco a while ago and they claim some traps need a higher temp bakeout. As I remember they said 340-360C ( check with them first!). Seems the packing can go to 400 but the glass wool packing would get activated.

Re: Moisture hump in voa run

Posted: Thu Nov 01, 2012 8:51 pm
by sllimbri65
Called Supelco and they said we could bake at 375 for 2 hours. Will try that and see if it helps.

We use graphite ferrules also. It has been changed a few times already. All leak test have passed.

Re: Moisture hump in voa run

Posted: Fri Apr 19, 2013 6:51 pm
by jenniferd
Hi sllimbri65,

I believe I have the same problem, how did you fix this?

Thanks!

Re: Moisture hump in voa run

Posted: Fri Apr 19, 2013 7:44 pm
by sllimbri65
It took a long time. We just kept running blanks and it went away by itself. Sorry, no magic bullet.

Re: Moisture hump in voa run

Posted: Fri Apr 19, 2013 7:52 pm
by jenniferd
OK, thank you!