switching off hplc's in end of the day
Posted: Fri Oct 26, 2012 12:20 pm
by chrom
Hi..
Is switching off the HPLC instruments is advisible in end off the day. Normally we keep the lamp switch off through software and remaining is in on mode. Insturments are connected through Empower software.
Thanks in advance.
Re: switching off hplc's in end of the day
Posted: Mon Oct 29, 2012 2:09 pm
by skunked_once
There is no fixed rule as to whether to switch the HPLC off at the end of the day. If I am going to use my instrument the next day, I leave it turned on but turn off the UV lamp. The only reason to turn off the instrument is if you need to restart to clear an error or lockup. I turn the instrument off over the weekend. Other users my have a different practice.
Posted: Mon Oct 29, 2012 2:25 pm
by danko
If the system/s in question, were privetly/personally owned and if the consumed electric power was charged to a privet person (not too often but just for the sake of the argument) I believe they’ll be switched off somewhat more frequently (perhaps on a daily basis) :wink:
Posted: Mon Oct 29, 2012 3:36 pm
by Johnny Rod
If you're using buffers in the mobile phase then you either need to flush the system out before turning it off, or leave it on low flow or recycle. Unless you really need to save on your electricity bill, most people leave them switched on when not in use so the column oven etc. is all equilibrated and ready to go.
Posted: Mon Oct 29, 2012 5:16 pm
by Vlad Orlovsky
we set up global shutdown at the end of the sequence if we running it after hours, in most cases this is done to preserve lifetime of the UV lamp.
Posted: Mon Oct 29, 2012 6:54 pm
by LC_labrat
Small or big lab, I have always re-booted HPLC instruments weekly.