-
- Posts: 5
- Joined: Fri Oct 28, 2011 2:19 pm
Advertisement
Smaller areas using hydrogen carrier compared to helium.
Discussions about GC and other "gas phase" separation techniques.
24 posts
Page 1 of 2
I injected the same sample with the same column flow and split ratio using hydrogen as the carrier and helium. My areas were noticeably less using the hydrogen. Has anyone seen this and what would be the cause?
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Golly, the answer depends upon which GC detector you are using. Do you know the detector you are using and can you tell us?
best wishes,
Rod
best wishes,
Rod
-
- Posts: 5
- Joined: Fri Oct 28, 2011 2:19 pm
I,m using a FID detector and helium as a make-up gas in both situations.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
You are changing the flame composition when you change the carrier to hydrogen.
This changes the response of the flame.
best wishes,
Rod
This changes the response of the flame.
best wishes,
Rod
-
- Posts: 5
- Joined: Fri Oct 28, 2011 2:19 pm
Thanks for the information.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
There are two main possibilities; a change in inlet discrimination due to lower pressures or a change in detector response due to more hydrogen going to the flame. Since for some reason you provide only the barest outline of your conditions it is anybody's guess which of these might apply. I would be surprised to find an impact on detector response unless you are using a megabore column.
Peter
Peter
Peter Apps
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
This is curious situation, as pretty much everyone who changes from helium carrier to hydrogen carrier does such to move away from helium cost or availablity. Using helium as the make-up gas only partially removes helium from the whole situation; most use nitrogen as make up gas when using hydrogen carrier.I injected the same sample with the same column flow and split ratio using hydrogen as the carrier and .... FID detector and helium as a make-up gas in both situations.
-
- Posts: 5
- Joined: Fri Oct 28, 2011 2:19 pm
Peter,
I was using a 15 meter megabore column.
I was using a 15 meter megabore column.
-
- Posts: 5
- Joined: Fri Oct 28, 2011 2:19 pm
Consumer Products Guy'
I understand that nitrogen would be the best choice as a make-up gas. We didn't have anything other than house nitrogen to do our test so I just replumbed the hydrogen to the carrier lines. Thanks.
I understand that nitrogen would be the best choice as a make-up gas. We didn't have anything other than house nitrogen to do our test so I just replumbed the hydrogen to the carrier lines. Thanks.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Guessing (again !) that your flow rate is quite high with such a short wide bore column - say 10 ml/min ? - then you have 10 ml/min extra hydrogen gong into the lfmae, and this is reducing its response. I expect that the baseline noise is also worse. What happens if you reduce the hydrogen setting on the FID gas contoller to restore the total hydrogen flow to the flame to what it was before ?Peter,
I was using a 15 meter megabore column.
Peter
Peter Apps
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
This was my guess-assumption too, Peter.
While a 0.2mm ID column would not change the flame much with a H2 carrier, the evidence that it had changed the flame suggested a 0.53 or 0.32mm ID column was being used.
Sometimes with so little information given it is difficult to produce a reasonable guess.
But we do try, don't we?
Even a change of H2 from 35cc/min to 37cc/min can make a difference. Certainly, 35 to 40cc/min can make a BIG difference, and doesn't always produce more FID noise, sometimes less, depending upon the makeup and the volume of air flow.
best wishes,
Rod
While a 0.2mm ID column would not change the flame much with a H2 carrier, the evidence that it had changed the flame suggested a 0.53 or 0.32mm ID column was being used.
Sometimes with so little information given it is difficult to produce a reasonable guess.
But we do try, don't we?
Even a change of H2 from 35cc/min to 37cc/min can make a difference. Certainly, 35 to 40cc/min can make a BIG difference, and doesn't always produce more FID noise, sometimes less, depending upon the makeup and the volume of air flow.
best wishes,
Rod
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Hi Rod
Great minds guess alike !!
Changing flows away from the generic "optimum" values can improve signal:noise from an FID, but I supsect that these days anyone who wants minimum lower detection limits would be using a selective detector.
One of the very big advantages of constant flow EPCs is that the flow of hydrogen carrier to an FID stays constant during temperature programming - if you run a megabore column at constant pressure with hydrogen carrier you get some spectacular baseline drift !
Peter
Great minds guess alike !!
Changing flows away from the generic "optimum" values can improve signal:noise from an FID, but I supsect that these days anyone who wants minimum lower detection limits would be using a selective detector.
One of the very big advantages of constant flow EPCs is that the flow of hydrogen carrier to an FID stays constant during temperature programming - if you run a megabore column at constant pressure with hydrogen carrier you get some spectacular baseline drift !
Peter
Peter Apps
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
You bring up a good point !
And I can only imagine the possibilities !
I am SO GLAD I did GC in the good ol' days of helium carrier. Something to tell my grandchildren (2 here now, and hopefully, more coming) someday.
best wishes,
Rod
And I can only imagine the possibilities !
I am SO GLAD I did GC in the good ol' days of helium carrier. Something to tell my grandchildren (2 here now, and hopefully, more coming) someday.
best wishes,
Rod
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Of course one of the benefits of constant head pressure (were talking the good really old days here) is that during temperature programming the declining volume flow of hydrogen carrier would cancel out some of the baseline drift from stationary phase bleed
Peter
Peter
Peter Apps
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Heavens !
You did not have a dual FID GC?
That is why columns were sold in pairs, so the bleed would be compensated by the parallel detector !
The amps always subtracted the bleed from the non-injected column, and you could switch A-B to B-A when a column got old or contaminated.
Oh, the days of scrubbing out the white powdery silicia deposits with a spatula! It brings tears to my eyes !
Rod
You did not have a dual FID GC?
That is why columns were sold in pairs, so the bleed would be compensated by the parallel detector !
The amps always subtracted the bleed from the non-injected column, and you could switch A-B to B-A when a column got old or contaminated.
Oh, the days of scrubbing out the white powdery silicia deposits with a spatula! It brings tears to my eyes !
Rod
24 posts
Page 1 of 2
Who is online
In total there are 26 users online :: 3 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Baidu [Spider] and 23 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Baidu [Spider] and 23 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science
