Problems to 2414 Refractive Index detector Waters

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi,
We have an GPC Waters Breeze System(Isocratic HPLC pump 1515; 717 plus Autosampler, 2414 Refractive Index detector and 3 columns Styragel HR2xHR 0.5xHR0.5).In this moment we changed 2 valves on the detector:Solenoid valve and Relief valve, because it flows in the same time on the purje mode and on the analysis.Two days after this the GPC system it works, but after it begining again .What's happening? Please help me.
Thank you,
Ileana
hello,

It is possible that you have too much pressure in the system. In this case, the liquid flows by purge and relief valve at the same time.
Please check if your waste tubing or your purge out tubing are not plugged.
Maybe the internal diameter of these tubes are to small. Try to work without tubing just for test my theory.

I hope you will find one solution quickly,

Pierre
Thank you for your answer.You have right.We changed the wasting tube and the system is ok.
Ileana
3 posts Page 1 of 1

Who is online

In total there are 38 users online :: 1 registered, 0 hidden and 37 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 37 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry