Choosing the right range

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I have a general question:
when I'm developping an HPLC method, I could find that the linearity will be in a very wide range of concentration.
My question is how can I choose the best range of lineartiy to work in routine.
Guidelines in general said that it must be neccesary to work between the 80% and 120% of the unknown concentration but, I know that if I work near the LQ the accuracy of my results will be worst than if I work in another range more concentrated.

Best regards

No single answer because sample prep will also affect results (e.g., if you have do to a dilution, that will add 0.5% or so to your CV).

I'd go with whatever minimizes sample workup; anything between about 5X LLOQ and whatever gives you about 0.5AU peak height should be OK.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Feedfetcher and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry