Evaporative light scattering detector
Posted: Thu Jun 23, 2005 2:59 pm
by JulesC
Dear members
Has anyone had experience with Light Scattering Detectors for carbohydrates? We bought a ELSD thinking it would be better for carbohydrates than RI detection (more stable).
The ELSD doesn't seem to be giving a linear response.
Any ideas/suggestions would be most welcome,
Best wishes
Julian
Posted: Thu Jun 23, 2005 3:21 pm
by Consumer Products Guy
We have ELSD. ELSD has a smaller linear range than does RI or UV detection. You may need a multi-point calibration.
We have ELSD. ELSD has a smaller linear range than does RI or UV detection. You may need a multi-point calibration. Some math wizard can probably provide scientific reasons.
Posted: Thu Jun 23, 2005 5:46 pm
by Kostas Petritis
Julian,
It is well known that the ELSD is not a direct linear response detector however linear calibration plots can be obtaned for several orders of magnitude by using double logrithnmic coordinates such as:
LogA=bLogC +loga
where C is your analyte concentration expressed in mgL-1 (for example) and A is the peak area response obtained after integration of the chromatographic peak of each analyte.
In other words quantitation and method validation is possible. For more information look at:
1) Mourey TH and Oppenheimer LE Anal Chem 1984 56, 2427-2434
2) Stolyhwo A et al J Chromatogr. 1984, 288, 253-275
3) Dreux M and Lafosse M, LC GC Int. 1996, 9, 148-156
4) Petritis et al. Chromatographia, 2004, 60, 293-298.
Posted: Thu Jun 23, 2005 8:06 pm
by John
You should be able to get stable enough results on the RI. If you are not getting stable results, your instrument/method may not be optimised.
I would choose RI over ELSD, for most cases. ELSD, though, has the advantages that it will work on a gradient and it truly is a universal detector.
J
Posted: Fri Jun 24, 2005 1:17 am
There are instances such as with nonionic surfactants that the ELSD is more sensitive (better S/N).