The issue is that the pressure needed to achieve required flow rates will be zero or lower if there is a vacuum at the outlet, helping to suck the carrier through the column. A restriction at the outlet (before the MS) is useful in some cases, though operators would often rather not add that complication.
Apparently, a restriction at the inlet end is even better, because then the whole column is at reduced pressure, which gives very high optimum flows, and faster separations. It is extra fiddle though, and ordinary EPCs struggle to work out the flows and pressures. Have a look on the Restek web site - Jack Cochran worked out some elegant ways of incorporating inlet restrictors into inlet liners.
Peter