ChemStation: target and qualifier ions m/z window

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
I'm setting up a GCMS method with SIM, and the ISTD is only a few m/z units away from the component of interest and has the same RT (it's a d-labeled variant). When I review the data, the software selects the wrong ion for the ISTD target ion; I'm guessing because the target ion of the unlabeled component is more abundant, and the m/z window is too large.

Where can I narrow the m/z window so that the software selects the correct target and qualifier ions?

We're using ChemStation E.02.02.
I think figured it out.

I opened the Extracted Ion Chromatogram window, and noticed that the amu window values at the bottom of the screen were +/- 4.568 (which happened to be the begin time of data acquisition after the 4.5 min solvent delay). I tried typing in the default values of -0.3 and +0.7 but they didn't change, and remained at 4.568. I tried 0, that didn't work. Then I hit the backspace key so the fields were empty, and presto! After reopening the window, the default values were shown and now used for my data analysis, so all the ions were correctly assigned.
2 posts Page 1 of 1

Who is online

In total there are 21 users online :: 2 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry