Molecular weight for Ethylene is 28, which would yield ions of 28 with m+1 of 29 and if you lose 1 hydrogen in ionization you get 27, cleave the double bond and you will see 14.
NItrogen(N2) mass is 28 (N) mass is 14.
You can never get a GC/MS totally leak free or your gasses totally pure without nitrogen impurities so you will always have a background for m/z 14 and 28 which would interfere with detection of Ethylene. FID would be much better for analyzing Ethylene as it would not have these interferences. Also some of the better columns for gas analysis are packed columns and those have too high flow rates for carrier gas to be used with the GC/MS without using a jet separator.
Thank you for your information. I was thinking that the low molecular weight of ethylene was a big issue, but never though of other aspects.
When talking about "jet separator", are you referring to the GC-MS interface which is used to interface the packed columns, rather that the direct coupling of the capillary column? Currently I only has a HP-5MS column. If I just lower the flow rate of the carrier gas (He), will it be a problem?