This molecule and the correspodning impurities where either one or two of the sulfate groups are removed. I do not have a scanning FL instrument anymore, so I cannot just measure it.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
The lower excitation λ you choose the greater risk for background quenching (i.e. the other things in the sample and even more importantly the mobile phase will absorb a greater deal of the incident energy/light as well as the fluorescence radiation /light). So, if both the initial product/analyte and its degradants exhibit reasonable absorbance at 263 nm it could be the best choice – all depending on the mobile phase. On the other hand – and especially if sensitivity is important, one should keep in mind that the lower wavelength you choose the higher energy is applied on the molecule thus more sensitivity can be achieved.Acc. to your suggestion, I should try to set the excitation wavelength to 226 nm and the emission wavelength to something in the range of 280-350 nm. Or should I go for the 263 nm absorption for excitation instead?