When carrying out annual calibration of HPLC systems (Waters 2695) is there any guidence from the FDA what should be covered, currently I carry out the below parameters

Detector
Wavelength Accuracy
Wavelength Linearity

Seperation Module
injector accuracy
Flow Rate
carosel positioning
Column heater Temperature
Valve Proportioning test

Does the 'noise and Drift' and the 'carry over' have to be checked on an annual basis as well. as in the initial installation of the HPLC

Thanks in advance
You can get away with a PQ in this situation. But you’ll have to describe it in an instruction.
The way I’ve done it previously, is as follows:
You pick up a challenging method of analysis – preferably one that makes full use of the system’s capabilities (fx. gradient elution, good system suitability tests, wavelength change etc.).
Write down in the PQ instruction that this will be your PQ and even more importantly the requirements that have to be met in order to approve the analytical run. Then describe the potential control chart or whatever. And finally write down that such a sequence would not be used for product release but exclusively for the purpose of controlling the equipment’s performance.
Then you are ready to run the standards, controls, blanks etc. that are relevant. And if the results are as described/expected then the system is performing according to your needs.

Best Regards
Learn Innovate and Share

Dancho Dikov
The FDA will expect you to carry out whatever system checks are called for in an SOP written to cover the subject. They offer no instrument specific guidance.

In our lab, we check injector precision and linearity. We also check pump mixing via step test and linear gradient formation. From the gradient formation run, we also calculate and note the system's dwell volume, though there is no spec. for this.

Wavelength linearity, I'm guessing is for DADs only?
Thanks,
DR
Image
Linearity is just different known percentage concentrations of the same solution, the absorbence is then measured at the same wavelength and the results are then used to calculate the sensitivity
Ah - I've done that, but referred to it as molar absorptivity (sp?).
Thanks,
DR
Image
5 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry