In regulated labs, when preparing Standards the primary standards are prepared in duplicate and compared. if the comparison is not acceptable (eg response difference > 2%) they are not used. If they compare with in the desired criteria on of these will be used to prepare the calibration.

From this data you can monitor the quality of STD production and also identify operator variations.

Having a wide variation in curve equations could be down to poor STD preps or an unstable method.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.
One question I have is what exactly is your variation from day to day? You said it was large but didn't give any information regarding how large. How many curves do you have and over what time span? Has any sort of pattern developed over time? Have there been any changes in your system, lab, equipment, procedure or personnel? It could be something as simple as a pipettor wearing or it could be more involved.
http://the-ghetto-chromatographer.blogspot.com/
3 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry