Hollow » Mon Oct 01, 2012 8:04 pm
hi
the alliance (2695) has no six-port valve or similar, so there can't be the problem.
The injector/sample loop is built somehow different.
The error message you've gotten guides you where to look first.
Seems that there is something wrong/plugged just near (or even inside) the accumulator. Otherwise I would expect just a "normal" high pressure shutdown.
To compensate the compressability of the solvents, the accumulator's controller measures the force needed to pump against the system pressure, which is sensed just after the accum outlet and compares theses two values.
If I remember correctly we had such an event once (unfortunately I can't remember the cause, yet).
To bring the pump to work again, I guess you need to shutdown the system and open the purge valve before you power on again, so that the piston can move and initialize itself without any restriction.
If error still exists, it's likely that there is a problem with the accumulator drive itself.
Maybe try to rebuild the piston's head and seals and inspect the plunger. Maybe there is some particles in there (debries from whatever... eg. GPV, primary seals...?). Maybe also have a look at the inline filter.