Question about Piracetam and its unstable peak.

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
It is a really big trouble. I have prepared every as the following:
Buffer: Na3PO4 0.02M adjust pH 7.0 with H3PO4.
Mobile phase: Buffer : CH3OH = 95:5

I run the HPLC with following conditions:
Flow rate: 0.5ml/min;
Wavelength: 220nm
C8 column 250x4.0
Run time: 20 minute

I use Buffer as Blank and to dissolve my sample.
I weight 250mg Piracetam to 50ml flask, dissolve with Buffer and sonicate in 1 minute then make up to volume with Buffer. After that , I draw 5ml and dilute to 50ml flask with the same solvent then filter it.
And here is my problem:
I often get 2 peaks: one is Piracetam (RT = 9.0) and another is Unknown Impurity (RT = 5.0). Peak due to Piracetam is very stable but the Unknown Impurity isn't even I filter 3 vials from only 1 flask, I always get the disparity in area (1.5 times or even double, triple). And the peak area even is unstable day by day. Sometimes is vey high (30000 - 35000) but sometimes is very low (4000 - 10000). I am sure that I did 100% correctly in procedure. I am looking forward your opinion and any idea helping me stabilize the Unk. Impurity peak area will be appreciated. Thanks for reading and helping!!!
I hope to get solution from you
My first question is:
what column temperature do you have? At pH7 it is critical to have a stable column temperature.
Please check.
Good luck.
Gerhard Kratz, Kratz_Gerhard@web.de
Sorry for not giving you enough information. I don't use column temperature. My room temperature is about 24 or 25 Celsius.
If my problem is column temperature. So, how many Celsius should I increase for my column temperature?
And one more problem, the buffer seem to be unstable after a few hours at room temperature. In the morning, I use buffer I have prepared to make my sample and get the Impurity's peak area is about 9000 - 14000. But in the evening, with the same buffer the peak area is always higher (20000-45000 sometimes even 80000). Today I tried to use the buffer after preparing as soon as possible and I will tell you the result tomorrow. Thanks for your help!
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 4 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider], John Guajardo, Semrush [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry