I need some help with method parameters for an Agilent 5975C MSD. I went to the OFN.sn.m in my software but it looks like someone change all the parameters. I am wanting the original factory parameters Thanks Drew.
-
- Posts: 75
- Joined: Fri Oct 07, 2011 10:16 am
From Agilent 5975 series operation manual page 76,
To verify system performance
Materials needed
• 1 pg/μL (0.001 ppm) OFN sample (5188-5348)
Verify the tune performance
1 Verify that the system has been pumping down f or at least 60minutes.
2 Set the GC oven temperature to 150 °C and the column flow to 1.0 mL/min.
3 In the Instrument Control view, select Checkout Tune from the Checkout
menu. The software will perform an autotune and print the report.
4 When the autotune has completed, save the method and then select
EvaluateTune from the Checkout menu.
The software will evaluate the last autotune and print a System Verification
– Tune report.
Verify the sensitivity performance
1 Set up t o inject 1 μL of OFN, either with the ALS or manually.
2 In the Instrument Control view, select Sensitivity Check from the Checkout
menu.
3 Click the appropriate icons in the Instrument. Edit window to edit the
method for the type of injection.
4 Click OK to run the method.
When the method is completed, an evaluation report will be printed.
Verify that rms signal-to-noise ratio meets the published specification.
Please see the Agilent Web site at www.agilent.com/chem for specifications.
To verify system performance
Materials needed
• 1 pg/μL (0.001 ppm) OFN sample (5188-5348)
Verify the tune performance
1 Verify that the system has been pumping down f or at least 60minutes.
2 Set the GC oven temperature to 150 °C and the column flow to 1.0 mL/min.
3 In the Instrument Control view, select Checkout Tune from the Checkout
menu. The software will perform an autotune and print the report.
4 When the autotune has completed, save the method and then select
EvaluateTune from the Checkout menu.
The software will evaluate the last autotune and print a System Verification
– Tune report.
Verify the sensitivity performance
1 Set up t o inject 1 μL of OFN, either with the ALS or manually.
2 In the Instrument Control view, select Sensitivity Check from the Checkout
menu.
3 Click the appropriate icons in the Instrument. Edit window to edit the
method for the type of injection.
4 Click OK to run the method.
When the method is completed, an evaluation report will be printed.
Verify that rms signal-to-noise ratio meets the published specification.
Please see the Agilent Web site at www.agilent.com/chem for specifications.
Jetjamnong
-
- Posts: 350
- Joined: Tue Jun 28, 2011 7:30 pm
Your software may have a backup of the original method - mine is in "C:\MSDChem\mssetup\5975". All the original installation stuff is in there.
3 posts
Page 1 of 1
Who is online
In total there are 56 users online :: 0 registered, 0 hidden and 56 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 56 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 56 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science