We have tried in 2 diferents CG (Agilent CG 6890) without success.
The method has been used for a long time now without problems, so we are running out of ideas.
Any idea??
Advertisement
Discussions about GC and other "gas phase" separation techniques.
What are your gas flow rates ? (FID incl. makeup, carrier)any idea why this could be happening??
We have tried in 2 diferents CG (Agilent CG 6890) without success.
The method has been used for a long time now without problems, so we are running out of ideas.
Any idea??
Our 68901's are not too old. Maybe 5-7 years or even less. They have a mantainance plan, with verifications twice a year. The conditions of the method should not be a problem since we have analised similar samples before with good results in the same GC.How old are your 6890s? at grad school i was working with old units (mid 90s), and the flow rate values that appear on the front panel didn't match the ones measured with a flow meter at the outlet of the FID, For example, for H2 i had it at 40 mL/min, but in reality it was 30 mL/min, same for air. I had a similar problem with an FID flame, that was solved by increasing my values of H2 and H20.
Now, in your particular case, I guess the MTBE solvent is adding extra oxygen to your flame. In that case, i'll probably increase my H2, and reduce my air flow, that may help, specially, if you indicate that it happens when the MTBE peak appears. by the way, what are your current values of H2 and H20 flow rates?
Gustavo/Supelco
Yes, sorry, I obviously meant uL.Let's assume that when you write "ml" you really mean "ul" for microlitres and go on from there.
You are doing a 1 ul splitless injection with a mystery flow rate into a mystery column at mystery temperatures, leaving us to guess wildly.
My wild guess is that your solvent is not what you think it is - did the onset of the problem coincide with opening a new bottle, or some decanting from stock into flasks or something similar ?
Peter
If column flow is 30 mL/min you may substantialy (!) reduce makeup flow or even turn it off completely.... Column flow 30mL/min (steady)...