Keeping chromatography condition same except detector, I am getting difference in impurity percentage ? When I run same sample on both detetctor I got different % impurity ? Is it possible or any reason for this?
For your information chromatographic condition given below
Column : Flexit C18 250x4.6, 5micron
Flow rate: 1.5ml/min
Mobile phase : Buffer:MeoH (80:20)
Detection at : 351nm
Detector : Waters 486 and Tosho UV8011
%Impurity : from waters 486 detector = 4.7%
%Impurity : from Tosho UV 8011 detector = 5.8%
All most of 1.0% difference.
I am doing % Area normalisation method for calculation.
For your information Only detector change other component is same.
Thanks in advance
Narendra