Our lab just bought a second-hand LC/MS (agilent 1100 LC with API 2000 MS). The technician from the vendor installed, tuned and gave us a basic trainning on it. However, when I started doing some test runs, I got very high basline noise (10^6, sometimes 10^7), and the major peaks are m/z 65, 83, 93, 97, 125, 129. After some digging, I found the m/z 65 may be caused by dimerization of methanol (2M+H), but I can not figure out what causes the problem and how to solve it. No one in our lab knows how to use MS. Therefore, I need help please!
Thanks
Mayber there is no real Problem?
I don’t know this system and cannot judge if the noise is normal or abnormal for this system and your conditions. But it is not uncommon to observe much noise with lc/ms especially when you start your scan-range with m/z 50 (many applications start with m/z 100). Therefore using background-subtraction is often done for lc/ms if you want to have nice mass spectras.
The mentioned dimerization of methanol is very common, depending on system, ionization-technique, tuning and used chemicals you observe often ion-cluster builded from you mobile phase (also Na+, K+; NH4+ is almost everywhere) and/or your analyte.
Maybe you should get a documented application for your system and try to reproduce it in your laboratory with your system.