Page 1 of 1
TPH - ultrasonic extraction
Posted: Wed Oct 03, 2012 10:20 pm
by pueblito
Dear Chromatographers.
We want to extract TPH from soils using the ultrasonic extraction method ( EPA 3550) intead of Soxhlet extraction especially because we have limited space available and for time issues, but the method does not include TPH extraction data, I would like to know if it´s possible to extract such analytes by ultrasonic extraction.
thanks a lot.
Re: TPH - ultrasonic extraction
Posted: Thu Oct 04, 2012 1:56 pm
by pueblito
Sorry.
The 3550 does not include TPH, but 8015 shows a pair of examples in tables 6 and 7 about ultrasonic extraction for Diesel.
bye.
Posted: Fri Oct 05, 2012 1:15 pm
by twranger
Certainly possible- would you be using gravimetric determination?
The problem may be regulatory- most (US) permits only use EPA 9071, the soxhlet method using n-hexane as solvent, since this is a "method defined analyte". Other solvents/extractions give different results.
Posted: Fri Oct 12, 2012 1:51 pm
by pueblito
Dear twranger.
Actually, the mehod we perfor is a GC-FID. The method 8015 include a pair of examples about Diesel ultrasonic extraction, I think we could use this technique for our soil extraction.
thanks for your help.