Hi all,

We have an HPLC system with only fluorescence detector. I am not sure if it can be used for analyzing caffeine.
Thanks for any advice.
Yes, you (we) can :wink:

Try Excitation 276 nm and Enission 380 nm.

Best regards
Learn Innovate and Share

Dancho Dikov
we use quinine available through agilent for FLD check
Thank you for answering,
With best regards.