we have a L-7400 UV-Detector Merck-Hitachi.I want to use its start-in function through an PE200 Series Pump or Autosampler using Relay-Box(D-line).Is there any specific cable for activating Contact Signal Input to activate Time-Program in the Detector. Because in the Instruction's Manual, they are referring D-Line Cable.
-
- Posts: 5
- Joined: Mon Nov 14, 2011 8:32 am
Hi
You need a special printcard insertet in the back of the detector, to get analog output, and then the e-DIO cable for start/stop signal (partnumber 890-6145).
BR
You need a special printcard insertet in the back of the detector, to get analog output, and then the e-DIO cable for start/stop signal (partnumber 890-6145).
BR
2 posts
Page 1 of 1
Who is online
In total there are 528 users online :: 0 registered, 0 hidden and 528 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 528 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 528 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science