Noisy baseline even without flow cell
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
Agilent 1200 with DAD. I got noisy baseline on it. First I tried stopped the pump, the noise kept the same level. Then, I pulled the flow cell out and found the same noisy baseline. Please help me and give some opinion or suggestions about it! Thank you in advance!
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
It's all about defining noise, or in other words, degree of zoom. If you zoom in enough you will see noise - no matter what. On the other hand, if you zoom out enough you will find the noisiest baseline completely plain 
To be more concrete: You might like to describe the issue/noise using numbers - absorbance units or Volts or whatever.
Best Regards
To be more concrete: You might like to describe the issue/noise using numbers - absorbance units or Volts or whatever.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 5
- Joined: Fri May 13, 2005 8:23 pm
Thanks for your reminding. The noise level is between 0.1 - 0.2 mAU (with setting 1 AU = 1 V). It's a little bit high when doing impurity profiling.
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
OK, not huge noise but not noise-free either.
The question now is whether it has been like this since the birth of the equipment you have or it’s something that happened over time – maybe last couple of weeks or something.
If it always has been like this you could try playing with the detector/software settings – i.e. Reduce the sampling rate and/or use the smoothing function.
If you’re used to see much less noise and it became worst recently (without you changing anything of importance), it could be the lamp – needs replacement.
Best Regards
The question now is whether it has been like this since the birth of the equipment you have or it’s something that happened over time – maybe last couple of weeks or something.
If it always has been like this you could try playing with the detector/software settings – i.e. Reduce the sampling rate and/or use the smoothing function.
If you’re used to see much less noise and it became worst recently (without you changing anything of importance), it could be the lamp – needs replacement.
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 5
- Joined: Fri May 13, 2005 8:23 pm
Thank you very much, danko!
5 posts
Page 1 of 1
Who is online
In total there are 23 users online :: 0 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 23 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 23 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science