connection problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
we have a L-7400 UV-Detector Merck-Hitachi.I want to use its start-in function through an PE200 Series Pump or Autosampler using Relay-Box(D-line).Is there any specific cable for activating Contact Signal Input to activate Time-Program in the Detector. Because in the Instruction's Manual, they are referring D-Line Cable.
Hi
You need a special printcard insertet in the back of the detector, to get analog output, and then the e-DIO cable for start/stop signal (partnumber 890-6145).

BR