Quattro Micro MRM fail
Posted: Mon Sep 24, 2012 4:55 pm
Hello everybody
Here's the new round of the fight: Q.M. vs me (Let's get ready to ruuuumble)
So what I faced
I was optimizing a method for mycotoxins determination. First I dealt with the ones which ionize in positive mode (14 compounds) and it seemed quite OK, ending with rather acceptable signal in MRM mode. Then I started with negative mode (7 compounds) and this is where the problems begun.
I performed the optimization by syringe infusion, then started with HPLC injection and got plain MRM chromatograms, no peaks at all! Only one of 7 compounds gave some peak only at giant concentration of 1 microgram per ml.
Then I tried the SIR mode - all 7 were present!
I made record of some separate in daughter scan mode (for example, compound DON has a precursor 355 and daughter 295 - then I record "Daughters of 355" , scan window 295+/-0.5) - again I get the intensive peak. But if I record MRM - no way!
After all, to exclude possible compound-dependant issues, I decided to make test on chloramphenicol (CAP), which normally inonizes well in negative mode. The same data - great intensity in SIR mode (321) , complete fail in MRM (321->152, 321->157).
I checked the mass calibration - it is OK, both mass accuracy and scan speed compensation.
Certainly the coliision gas is on and the pressure is enough. The collision energy shows the appropriate readback.
Now there's a lack of ideas, I just get lost.
Later I'm gonna attach the chromnatograms and screenshots of Tune page or MS Method window, if that will be of use for anyone who has any ideas.
Here's the new round of the fight: Q.M. vs me (Let's get ready to ruuuumble)
So what I faced
I was optimizing a method for mycotoxins determination. First I dealt with the ones which ionize in positive mode (14 compounds) and it seemed quite OK, ending with rather acceptable signal in MRM mode. Then I started with negative mode (7 compounds) and this is where the problems begun.
I performed the optimization by syringe infusion, then started with HPLC injection and got plain MRM chromatograms, no peaks at all! Only one of 7 compounds gave some peak only at giant concentration of 1 microgram per ml.
Then I tried the SIR mode - all 7 were present!
I made record of some separate in daughter scan mode (for example, compound DON has a precursor 355 and daughter 295 - then I record "Daughters of 355" , scan window 295+/-0.5) - again I get the intensive peak. But if I record MRM - no way!
After all, to exclude possible compound-dependant issues, I decided to make test on chloramphenicol (CAP), which normally inonizes well in negative mode. The same data - great intensity in SIR mode (321) , complete fail in MRM (321->152, 321->157).
I checked the mass calibration - it is OK, both mass accuracy and scan speed compensation.
Certainly the coliision gas is on and the pressure is enough. The collision energy shows the appropriate readback.
Now there's a lack of ideas, I just get lost.
Later I'm gonna attach the chromnatograms and screenshots of Tune page or MS Method window, if that will be of use for anyone who has any ideas.