Here's a new one I got stuck with:
When I increased the desolvation gas flow up to 200 L/h and more, the Capillary Voltage and Cone Voltage turned off. With lower gas flow it was OK.
I couldn't see any logical reasons except for gas quality or pressure failures.
The pressure given by generator is appropriate for QM (>6 bar).
I thought the possible moisture in nitrogen could be a reason. I opened the N2 generator, checked the filters, indeed there was some water (I think - the auto-drain failed). I cleared away off the water and let the N2 generator to pump into atmosphere within 2 days (in order to pump out the water residues if there were any).
After that the QM let me to increase desolvation gas up to 400 L/h. If higher - again the Capillary Voltage and Cone Voltage turn off.
To make things clear about gas quality, I'm gonna attach a pure N2 cylinder instead of generator.
But I'm not completely sure that it is all about moisture (because I did my best to get rid of it).
Maybe anyone dealt with such issue?