Flow Rate
Posted: Tue Jun 14, 2005 2:56 pm
by carnie
I measure the flow before entering the column and it's 1 ml/min. I connect the column and the flow exiting the column is about 0.6 ml/min. I thought that the column may have become blocked but the back pressure has dropped where I would have expected an increase. Any ideas?
flow rate
Posted: Tue Jun 14, 2005 3:40 pm
by skunked_once
I suspect that you have a check valve problem. There is very little back pressure without the column connected so the check valves don't have to hold much pressure. When you connect the column, the back pressure increases significantly and the check valves may not hold the pressure.
Try replacing them with new check valves.
Posted: Tue Jun 14, 2005 3:54 pm
by Mark Tracy
While you are replacing checkvalves, replace the piston seals and inspect the pistons for damage.
Posted: Wed Jun 15, 2005 7:15 am
by HW Mueller
A good argument for monitoring flow, rather than completely relying on the pump setting.
Posted: Fri Jun 17, 2005 1:20 am
by carnie
I released that it wasn't a pump problem by switching pumps. However, I discovered a leak in the autosampler injection valve. Problem solved. Thanks for all your sugestions which helped me look in the right direction.