First, the peak is chromatographing unsymmetrical.
Second, if you break the peak down in to two sections and get the area response, A + B doesn't = C. When quantitating the unsymmetrical part of the peak i get a response of 8800 and when i quantitate the "normal looking" part of the peak I get 312000. But when I quant the two parts together the response is 510000.
I have changed the entire P&T (including transfer lines), column, inlet liners, adjusted split ratios from 20:1 to 40:1 and reinstalled software.
I believe this has to do with the CO2 peak, since that peak is chromatographing strangely also.
When I direct inject I don't have the problem with DCDFM but when the air comes out after the gaseous compounds, while much smaller than in a normal purged standard, it gives a unsymmetrical peak.
The issue is rather difficult to explain but i would be more than happy to email a file of the standard that i ran today if anyone would like to see.
Thanks in advance
Chris