Where does these noise come from?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
I try to setup a Water's HPLC, but see noise about 0.05 AU from 0 ~ 2.8 min, after 2.8 min, the noise is gone (UV absorption down to 0.00005 AU). I try different mobile phase, and clean the system overnight, but the noise is still there. Could anybody please give some hint where the noise came from? Thanks
-
- Posts: 38
- Joined: Fri Aug 24, 2012 1:41 pm
Is this isocratic? Does this happen when you run the analysis with no injection?
-
- Posts: 7
- Joined: Tue Nov 08, 2011 6:43 pm
What I have tried uptil now:
I remove the column, replace with a union, isocratic (H2O:ACN 50:50), tried 2 flow speed, 1.0 ml/min and 1.5 ml/min,
and still repeating the same result: 0 - 2.8 min, I have relatively big noise (~ 0.05 AU), right at 2.8 min, the signal drop down to 0.0005 AU, looks like a thin line.
I will try the monitoring mode (is this what you mean without injection) next,
I also feel this is something related with injection, so I try to change the injection related parameters, so far I tried change the needle wash time (from normal to double), but result remain the same.
I remove the column, replace with a union, isocratic (H2O:ACN 50:50), tried 2 flow speed, 1.0 ml/min and 1.5 ml/min,
and still repeating the same result: 0 - 2.8 min, I have relatively big noise (~ 0.05 AU), right at 2.8 min, the signal drop down to 0.0005 AU, looks like a thin line.
I will try the monitoring mode (is this what you mean without injection) next,
I also feel this is something related with injection, so I try to change the injection related parameters, so far I tried change the needle wash time (from normal to double), but result remain the same.
Is this isocratic? Does this happen when you run the analysis with no injection?
-
- Posts: 2174
- Joined: Tue Aug 17, 2004 7:59 pm
Major mis-match between rinse solvent and mobile phase?
Thanks,
DR

DR

-
- Posts: 392
- Joined: Thu Sep 08, 2005 4:12 am
did your detector have an event set from 0-2.8min, which is strange though?
Excel
-
- Posts: 38
- Joined: Fri Aug 24, 2012 1:41 pm
Are you seeing this when you don't make an injection? Run an isocratic separation and inject the mobile phase. See if you see the disturbance. Try injecting 0, 10, and 20 uL. Are you seeing any differences?Major mis-match between rinse solvent and mobile phase?
6 posts
Page 1 of 1
Who is online
In total there are 345 users online :: 0 registered, 0 hidden and 345 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 345 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 345 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science