Varian 3800 Saturn 2000 MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Dear All,
Does anyone know - is it possible to use hydrogen in this system as a replacement for helium?
Regards
WK
I'm Sorry I Haven't A Clue - Just A Minute - The Unbelievable Truth
I'm using hydrogen for our ITS40s, Magnums and Saturn IIs so it should work for you too.

I just recommend to vent the MS with a dry inert gas like nitrogen or argon.
Hi Hbj,
Thanks for the reply. How do you "vent" the trap with inert gas?
Add a small flow through the CI line??
Regards
WK
I'm Sorry I Haven't A Clue - Just A Minute - The Unbelievable Truth
I meant venting to atmospheric pressure in case of an emergency or scheduled downtime for maintenance.

Venting with inert gas eliminates the theoretical possibility of a hydrogen explosion that might happen when air is used as vent gas. In fact, Finnigan did require the customer to connect the vent line to either nitrogen or helium. Otherwise, no on-site support was provided.
4 posts Page 1 of 1

Who is online

In total there are 105 users online :: 2 registered, 0 hidden and 103 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Majestic-12 [Bot] and 103 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry