I am purchasing a 250 * 4.6 mm (5um) analytical SAX column and have multiple choices as to what ID i can get the guard column in. Price wise, the 20 * 4 mm option is the best, compared to the best ID match of 17*4.6 or even 40*4.6 mm guards.
Flow rate: 1ml/min
Mobile: PO4 buffer with MeCN, gradient

I would apprciate any comments on the pros and cons, if any, of choosing a guard with a smaller ID than the analytical column.

i would imagine that with the 4 mm rather than 4.6 mm ID guard, backpressure may be a little higher, but of course with the guard being very short, this increase would be minimal. Sample plug i would imagine would be 'tighter' with the smaller ID guard.....

thanks for any comments :D

Greg
From the point of view of chromatographic performance, it probably doesn't matter very much, with a slight edge toward the smaller id as representing a smaller fraction of the total stationary phase volume. If your samples are very dirty, the larger id will last longer.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thanks Tom, as i expected, but nice to hear some affirmation.

BTW, we meet at the HPLC troubleshooting course at HPLC 2012 Anaheim, I'm the Aussie guy!

cheers
Greg
3 posts Page 1 of 1

Who is online

In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry