mlumanog,
Several comments. First, try removing your column from the detector and see if you get bubbles in a vial of methanol or isopropanol. You may not have any flow. If you do have flow, you can try squeezing the tubing from the flow meter around the end of the column to measure flow. Second, you don't say what your column head pressure is on the front gauge. That would be good to know. Third, you should tell us what column you are using since the id of the column will dictate the desired flow rate to get the optimum linear velocity (and the proper pressure regime.) Fourth, if you set your column flow to 2-3 mL/min (not bad for a 0.32) then you will need 100 - 150 out the split vent to make your desired 50:1. That's a lot helium.....
The way the 5890 works, you set the total flow high so that you can get a desired back pressure on the gauge, measure the column flow and adjust with column head pressure knob, then when you are happy with that you wind down the total flow to get the desired split flow. Use the two set flows to get ratio at this point.
Now the puzzler. You opened the aux gas and found the split vent at 70 ml/min. They should have nothing to do with each other so if you close the aux gas you should still have flow out the split vent. Is that the case?
Best regards,
AICMM
Sorry. My column head pressure is at 12. The column is a restek capillary column, 30m, 0.32 mb.
When I closed the aux valve and waited for a while, there was no flow out of the split inlet vent. At least none I could tell with my bubble meter.
Thanks for the tips. I'll try checking the detector end with isopropanol to check for any flow at all.