GC-MS inlet leak
Posted: Thu Sep 13, 2012 12:38 am
Hello,
I am using an Agilent 6890N GC connected to an MS. I am having trouble getting the vacuum to purge the MS chamber of atmosphere because of a leak in the inlet (i think...). If I remove the column from the inlet and plug the end of it with one of those blue rubber things that comes with new columns, then I can "air and water test" and the chamber gets emptied and the m/z = 69 is as normal. As soon as I connect the column back into the inlet (cool on column), the chamber is filled with atmosphere and m/z = 69 is not visible on the spectrum, even if I let it pump down overnight.
I have replaced ferrules (85/15 vespel graphite), replaced the septum, cut the column, and cleaned all the surfaces involved (where the ferrule makes contact with the bottom of the inlet, for example). No luck.
Does anyone have any suggestions of what to try next? Unfortunately I cannot afford to pay an Agilent rep to come out to our site and help us...
Thanks in advance,
Luke
I am using an Agilent 6890N GC connected to an MS. I am having trouble getting the vacuum to purge the MS chamber of atmosphere because of a leak in the inlet (i think...). If I remove the column from the inlet and plug the end of it with one of those blue rubber things that comes with new columns, then I can "air and water test" and the chamber gets emptied and the m/z = 69 is as normal. As soon as I connect the column back into the inlet (cool on column), the chamber is filled with atmosphere and m/z = 69 is not visible on the spectrum, even if I let it pump down overnight.
I have replaced ferrules (85/15 vespel graphite), replaced the septum, cut the column, and cleaned all the surfaces involved (where the ferrule makes contact with the bottom of the inlet, for example). No luck.
Does anyone have any suggestions of what to try next? Unfortunately I cannot afford to pay an Agilent rep to come out to our site and help us...
Thanks in advance,
Luke